Monitoring the bone degradation-induced loosening of osseointegrated percutaneous implant using vibration analysis

Monitoring the bone degradation-induced loosening of osseointegrated percutaneous implant using vibration analysis

Qingsong Zhou, Benjamin Steven Vien, L.R. Francis Rose, M.K. Russ, M. Fitzgerald, Wing Kong Chiu

download PDF

Abstract. Osseointegrated implants attach prosthetic devices directly to the skeletal system with a stable connection, allowing amputees to control prostheses with good proprioception. However, periprosthetic cortical thinning can occur if the remaining bone is stress shielded after implantation, resulting in aseptic implant loosening, and requiring revision surgery. It is therefore important to quantitatively monitor the degree of bone loss and probability of loosening to provide evidence for doctors to plan treatment. This computational study investigates the vibration behavior of a bone-implant construct by monitoring its transient response over varying degrees of bone degradation. The thickness of distal bone is progressively reduced through a revolving material removal to simulate cortical thinning due to stress shielding. First, bonded contact is adopted to represent the case of the fully osseointegrated implant under bone resorption, leading to a linear vibrational response. Next, frictional contact with 0.05 mm over-fit offset is adopted to include the non-linear contact behavior when loosening is induced by cortical thinning. The torsional mode shape of the bone-implant construct is identified through cross-spectrum analysis. The results suggest that the change in the natural frequency of the first torsional mode provides the most sensitive indicator of cortical thinning and implant loosening. The findings underpin the potential of vibration analysis in the monitoring of bone degradation-induced implant looseness.

Keywords
Bone Resorption, Vibration Analysis, Finite Element Modelling, Transfemoral Osseointegrated Prosthesis

Published online 3/30/2023, 10 pages
Copyright © 2023 by the author(s)
Published under license by Materials Research Forum LLC., Millersville PA, USA

Citation: Qingsong Zhou, Benjamin Steven Vien, L.R. Francis Rose, M.K. Russ, M. Fitzgerald, Wing Kong Chiu, Monitoring the bone degradation-induced loosening of osseointegrated percutaneous implant using vibration analysis, Materials Research Proceedings, Vol. 27, pp 32-41, 2023

DOI: https://doi.org/10.21741/9781644902455-5

The article was published as article 5 of the book Structural Health Monitoring

Content from this work may be used under the terms of the Creative Commons Attribution 3.0 license. Any further distribution of this work must maintain attribution to the author(s) and the title of the work, journal citation and DOI.

References
[1] Newcombe, L., et al., Effect of amputation level on the stress transferred to the femur by an artificial limb directly attached to the bone. Medical Engineering & Physics, 2013. 35(12): p. 1744-1753. https://doi.org/10.1016/j.medengphy.2013.07.007
[2] Hagberg, K. and R. Brånemark, One hundred patients treated with osseointegrated transfemoral amputation prostheses–rehabilitation perspective. Journal of Rehabilitation Research & Development, 2009. 46(3).
[3] Overmann, A. and J. Forsberg, The state of the art of osseointegration for limb prosthesis. Biomedical engineering letters, 2020. 10(1): p. 5-16. https://doi.org/10.1007/s13534-019-00133-9
[4] Brånemark, R., et al., Osseointegration in skeletal reconstruction and rehabilitation. J Rehabil Res Dev, 2001. 38(2): p. 1-4.
[5] Hagberg, K., et al., Osseoperception and osseointegrated prosthetic limbs, in Psychoprosthetics. 2008, Springer. p. 131-140. https://doi.org/10.1007/978-1-84628-980-4_10
[6] Gupta, S. and K.J. Loh, Monitoring osseointegrated prosthesis loosening and fracture using electrical capacitance tomography. Biomedical engineering letters, 2018. 8(3): p. 291-300. https://doi.org/10.1007/s13534-018-0073-4
[7] Gromov, K., et al., Do rerevision rates differ after first-time revision of primary THA with a cemented and cementless femoral component? Clinical Orthopaedics and Related Research®, 2015. 473(11): p. 3391-3398. https://doi.org/10.1007/s11999-015-4245-6
[8] Belgica, A.O., Stress shielding and bone resorption in THA: clinical versus computer-simulation studies. Acta Orthopaedica Belgica, 1993. 59(1): p. 118-129.
[9] Huiskes, R., H. Weinans, and B. Van Rietbergen, The relationship between stress shielding and bone resorption around total hip stems and the effects of flexible materials. Clinical orthopaedics and related research, 1992: p. 124-134. https://doi.org/10.1097/00003086-199201000-00014
[10] Frost, H.M., A 2003 update of bone physiology and Wolff’s Law for clinicians. The Angle Orthodontist, 2004. 74(1): p. 3-15.
[11] Gefen, A., Computational simulations of stress shielding and bone resorption around existing and computer-designed orthopaedic screws. Medical and Biological Engineering and Computing, 2002. 40(3): p. 311-322. https://doi.org/10.1007/BF02344213
[12] Huiskes, R., The various stress patterns of press-fit, ingrown, and cemented femoral stems. Clin Orthop, 1990. 2612738. https://doi.org/10.1097/00003086-199012000-00006
[13] Bragdon, C.R., et al., Differences in stiffness of the interface between a cementless porous implant and cancellous bone in vivo in dogs due to varying amounts of implant motion. The Journal of arthroplasty, 1996. 11(8): p. 945-951. https://doi.org/10.1016/S0883-5403(96)80136-7
[14] Caouette, C., L.H. Yahia, and M. Bureau, Reduced stress shielding with limited micromotions using a carbon fibre composite biomimetic hip stem: a finite element model. Proceedings of the Institution of Mechanical Engineers, Part H: Journal of Engineering in Medicine, 2011. 225(9): p. 907-919. https://doi.org/10.1177/0954411911412465
[15] Olufsen, S.N., Numerical Analysis of Primary Stability on Cementless Hip Prostheses. 2012, Institutt for konstruksjonsteknikk.
[16] Mirulla, A.I., et al., Biomechanical analysis of two types of osseointegrated transfemoral prosthesis. Applied Sciences, 2020. 10(22): p. 8263. https://doi.org/10.3390/app10228263
[17] Nebergall, A., et al., Stable fixation of an osseointegated implant system for above-the-knee amputees: titel RSA and radiographic evaluation of migration and bone remodeling in 55 cases. Acta orthopaedica, 2012. 83(2): p. 121-128. https://doi.org/10.3109/17453674.2012.678799
[18] Cachão, J.H., et al., Altering the Course of Technologies to Monitor Loosening States of Endoprosthetic Implants. Sensors, 2020. 20(1): p. 104. https://doi.org/10.3390/s20010104
[19] Shao, F., et al., Natural frequency analysis of osseointegration for trans-femoral implant. Annals of Biomedical Engineering, 2007. 35(5): p. 817-824. https://doi.org/10.1007/s10439-007-9276-z
[20] Cairns, N.J., et al., Ability of modal analysis to detect osseointegration of implants in transfemoral amputees: a physical model study. Medical & biological engineering & computing, 2013. 51(1): p. 39-47. https://doi.org/10.1007/s11517-012-0962-0
[21] Ruther, C., et al. A new approach for diagnostic investigation of total hip replacement loosening. in International Joint Conference on Biomedical Engineering Systems and Technologies. 2011. Springer.
[22] Rieger, J.S., et al., A vibrational technique for diagnosing loosened total hip endoprostheses: An experimental sawbone study. Medical Engineering & Physics, 2013. 35(3): p. 329-337. https://doi.org/10.1016/j.medengphy.2012.05.007
[23] Lannocca, M., et al., Intra-operative evaluation of cementless hip implant stability: A prototype device based on vibration analysis. Medical engineering & physics, 2007. 29(8): p. 886-894. https://doi.org/10.1016/j.medengphy.2006.09.011
[24] Chiu, W.K., et al., Healing assessment of fractured femur treated with an intramedullary nail. Structural Health Monitoring, 2021. 20(3): p. 782-790. https://doi.org/10.1177/1475921718816781
[25] Vien, B.S., et al., A vibration analysis strategy for quantitative fracture healing assessment of an internally fixated femur with mass-loading effect of soft tissue. Structural Health Monitoring, 2021. 20(6): p. 2993-3006. https://doi.org/10.1177/1475921720978224
[26] Qi, G., W.P. Mouchon, and T.E. Tan, How much can a vibrational diagnostic tool reveal in total hip arthroplasty loosening? Clinical Biomechanics, 2003. 18(5): p. 444-458. https://doi.org/10.1016/S0268-0033(03)00051-2
[27] Georgiou, A. and J. Cunningham, Accurate diagnosis of hip prosthesis loosening using a vibrational technique. Clinical Biomechanics, 2001. 16(4): p. 315-323. https://doi.org/10.1016/S0268-0033(01)00002-X
[28] Alshuhri, A.A., et al., Non-invasive vibrometry-based diagnostic detection of acetabular cup loosening in total hip replacement (THR). Medical Engineering & Physics, 2017. 48: p. 188-195. https://doi.org/10.1016/j.medengphy.2017.06.037
[29] Tomaszewski, P., et al., A comparative finite-element analysis of bone failure and load transfer of osseointegrated prostheses fixations. Annals of biomedical engineering, 2010. 38(7): p. 2418-2427. https://doi.org/10.1007/s10439-010-9966-9
[30] Thesleff, A., et al., Biomechanical characterisation of bone-anchored implant systems for amputation limb prostheses: a systematic review. Annals of biomedical engineering, 2018. 46(3): p. 377-391. https://doi.org/10.1007/s10439-017-1976-4
[31] Reif, T.J., et al., Early experience with femoral and tibial bone-anchored osseointegration prostheses. JBJS Open Access, 2021. 6(3). https://doi.org/10.2106/JBJS.OA.21.00072
[32] Lu, S., et al., Experimental Investigation of Vibration Analysis on Implant Stability for a Novel Implant Design. Sensors, 2022. 22(4): p. 1685. https://doi.org/10.3390/s22041685
[33] Weinans, H., R. Huiskes, and H. Grootenboer, Effects of fit and bonding characteristics of femoral stems on adaptive bone remodeling. 1994. https://doi.org/10.1115/1.2895789
[34] Frölke, J., R. Leijendekkers, and H. Van de Meent, Osseointegrated prosthesis for patients with an amputation. Der Unfallchirurg, 2017. 120(4): p. 293-299. https://doi.org/10.1007/s00113-016-0302-1
[35] Viceconti, M., et al., Large-sliding contact elements accurately predict levels of bone–implant micromotion relevant to osseointegration. Journal of biomechanics, 2000. 33(12): p. 1611-1618. https://doi.org/10.1016/S0021-9290(00)00140-8
[36] Prochor, P. and E. Sajewicz, The influence of geometry of implants for direct skeletal attachment of limb prosthesis on rehabilitation program and stress-shielding intensity. BioMed Research International, 2019. 2019. https://doi.org/10.1155/2019/6067952
[37] Kerschen, G., et al., Nonlinear normal modes, Part I: A useful framework for the structural dynamicist. Mechanical systems and signal processing, 2009. 23(1): p. 170-194. https://doi.org/10.1016/j.ymssp.2008.04.002
[38] Balasubramanian, P., et al., Nonlinear vibrations of beams with bilinear hysteresis at supports: interpretation of experimental results. Journal of Sound and Vibration, 2021. 499: p. 115998. https://doi.org/10.1016/j.jsv.2021.115998
[39] Lee, W.C., et al., Kinetics of transfemoral amputees with osseointegrated fixation performing common activities of daily living. Clinical biomechanics, 2007. 22(6): p. 665-673. https://doi.org/10.1016/j.clinbiomech.2007.02.005